On Nov 8, 11:22=A0am, Tim Wescott <t...@seemywebsite.com> wrote:
> On Tue, 08 Nov 2011 10:08:48 -0700, Jim Thompson wrote: > > On Tue, 08 Nov 2011 10:18:14 -0600, Tim Wescott <t...@seemywebsite.com> > > wrote: > > >>On Tue, 08 Nov 2011 06:41:54 -0800, halong wrote: > > >>> Hi all, > > >>> I have a 200 MHz clock signal to send across the barrier of 5 KV I'm > >>> thinking of some RF transformer - perhaps "air core" ... but can't > >>> find any good candidate > > >>Have you tried widening your search to just specify a 200MHz signal > >>frequency and galvanic isolation? =A0I think the 200MHz part isn't a bi=
g
> >>deal, but the galvanic isolation may be. > > >>If EMI is a concern then an air core transformer may not be a good > >>choice: it's going to leak RF like a sieve, and the spacing you'll need > >>for 5k isolation won't help. > > >>If your production volumes are low, anything that works right off the > >>bat is way less expensive than you think -- engineering time costs > >>money, after all. =A0So you may want to at least price out your fiber > >>optics solution. > > > Fiber optics? =A0Definitely best for a clock! > > I hadn't mentioned anything, but presumably he wants to get data back > from wherever his clock is going -- fiber would be good for that, too. > > --www.wescottdesign.com
You bet :-) The clock has the timing reference information, that's why we need it...otherwise we can populate a crystal on the other side :-) Hmmm, speaking out of this I used to dream about a wireless clock reference, and it's available everywhere in a closed system :-)))))
On Tuesday, November 8, 2011 6:41:54 AM UTC-8, halong wrote:

> I have a 200 MHz clock signal to send across the barrier of 5 KV > I'm thinking of some RF transformer - perhaps "air core" ... but can't > find any good candidate
You can get some optoisolators with 5 kV ratings, but not 200 MHz. If, however, it's a clock you're concerned with, can you divide down the clock to lower frequency, send the lower frequency through a standard optoisolator, and then reconstruct with a PLL on the HV side? It's not hard to build transformers for RF, but your application - 5 kV and 200 MHz- is unusual enough that I'd not expect to find any on-the-shelf solution. You'll have to wind one bobbin with HV insulated wire, and probably ground-shield between that bobbin and the core, then put the LV winding on a different bobbin, then assemble two C cores, maintaining bobbin spacing.
On Nov 8, 11:41=A0am, whit3rd <whit...@gmail.com> wrote:
> On Tuesday, November 8, 2011 6:41:54 AM UTC-8, halong wrote: > > I have a 200 MHz clock signal to send across the barrier of 5 KV > > I'm thinking of some RF transformer - perhaps "air core" ... but can't > > find any good candidate > > You can get some optoisolators with 5 kV ratings, but not 200 MHz. > If, however, it's a clock you're concerned with, can you divide down > the clock to lower frequency, send the lower frequency through > a standard optoisolator, and then reconstruct with a PLL on the HV > side? > > It's not hard to build transformers for RF, but your application - > 5 kV and 200 MHz- is unusual enough that I'd not expect to > find any on-the-shelf solution. =A0 You'll have to wind one bobbin > with HV insulated wire, and probably ground-shield between that bobbin > and the core, then put the LV winding on a different > bobbin, then assemble two C cores, maintaining bobbin spacing.
Hi whit3rd, Thanks for suggestion, divide to lower frequency then multiply it back to overcome the 200MHz is a good idea... What my concern about is the distortion/ jitter we adding up when the clock going through such stages...
halong wrote:
> Hi all, > > I have a 200 MHz clock signal to send across the barrier of 5 KV > I'm thinking of some RF transformer - perhaps "air core" ... but can't > find any good candidate >
Medical, defibrillator-proof? That's pretty much my home turf. I often use ferrite toroids, mostly #43 material and sometimes #61. Then 2-3 turns, bifilar winding. Bifilar is key here, without that it's not going to work. One of the wires needs to be at least MIL-STD 22759 amendment 1 (5kV tested). Aircraft suppliers have that. Then the usual, strict training, assembly and storage requirements, all ECO'd, and yada yada. If it's done right you can see amazing bandwidths. I did one interface (defibrillator-proof) where they needed 8nsec pulses and the slopes were less than 2nsec. Pretty much limited by what the driver circuitry could provide. "Air core" becomes feasible at that frequency. You could possibly have loops on both sides of the board. But the compliance folks might have a hissy fit it you can't furnish the required documentation for the circuit board core material. A serious downside of air coupling is ratiated emissions which can hit you from behind during EMC cert. -- Regards, Joerg http://www.analogconsultants.com/
On Tue, 8 Nov 2011 09:19:41 -0800 (PST), halong <ccon67@netscape.net>
wrote:

>On Nov 8, 10:04&#4294967295;am, John Larkin ><jjlar...@highNOTlandTHIStechnologyPART.com> wrote: >> On Tue, 8 Nov 2011 07:29:40 -0800 (PST), halong <cco...@netscape.net> >> wrote: >> >> >> >> >On Nov 8, 9:09&#4294967295;am, John Larkin >> ><jjlar...@highNOTlandTHIStechnologyPART.com> wrote: >> >> On Tue, 8 Nov 2011 06:41:54 -0800 (PST), halong <cco...@netscape.net> >> >> wrote: >> >> >> >Hi all, >> >> >> >I have a 200 MHz clock signal to send across the barrier of 5 KV >> >> >I'm thinking of some RF transformer - perhaps "air core" ... but can't >> >> >find any good candidate >> >> >> Fiber optics! >> >> >> Or capacitively couple into a tuned circuit. This is RF! >> >> >> John >> >> >Hi Jan, it's medical environment where any extra EM is regulated.., >> >few winding coils may work better than the dipoles >> >> >John: capacitive coupling... will it add more crying in hospital ? >> >> >Optics sounds feasible...but cost more than the air core transformer >> >for sure... >> >> What's the distance and geometry? >> >> Does phase shift matter? >> >> John > >Phase shift is no problem. The clock needs to cross the two "islands" >on same PCB...5KV is the different voltage... > >btw, I think this would be good candidate > >http://www.analog.com/en/interface/digital-isolators/adum2400/products/product.html > >but what I worry about is the pulse with distortion, the datasheet >rates max distortion = 2ns... Does that means either reference clock- >edge (rise/fall) will be 2 ns jitter ?...
That part is only rated to 400 VRMS continuous, but it is OK at 5KV RMS for short-term use. But it's not fast enough to pass a 200 MHz clock. Some sort of capacitive or inductive coupling would probably be best, with a tuned circuit and comparator in the receiver. It would be fun to use the pcb itself as the capacitors or as the transformer, with the entire thickness of the board as the insulator. John
On Tue, 08 Nov 2011 10:55:39 -0800, Joerg <invalid@invalid.invalid>
wrote:

>halong wrote: >> Hi all, >> >> I have a 200 MHz clock signal to send across the barrier of 5 KV >> I'm thinking of some RF transformer - perhaps "air core" ... but can't >> find any good candidate >> > >Medical, defibrillator-proof? That's pretty much my home turf. I often >use ferrite toroids, mostly #43 material and sometimes #61. Then 2-3 >turns, bifilar winding. Bifilar is key here, without that it's not going >to work. >
For fixed frequency, you wouldn't need bifilar. Just tune both sides and live with a lot of leakage inductance. A few turns of HV wire on opposite sides of a little Micrometals toroid maybe. How about a couple of Mini-Circuits RF transformers, balanced on both ends, connected with a pair of low capacitance, high-voltage ceramic caps? John
John Larkin wrote:
> On Tue, 08 Nov 2011 10:55:39 -0800, Joerg <invalid@invalid.invalid> > wrote: > >> halong wrote: >>> Hi all, >>> >>> I have a 200 MHz clock signal to send across the barrier of 5 KV >>> I'm thinking of some RF transformer - perhaps "air core" ... but can't >>> find any good candidate >>> >> Medical, defibrillator-proof? That's pretty much my home turf. I often >> use ferrite toroids, mostly #43 material and sometimes #61. Then 2-3 >> turns, bifilar winding. Bifilar is key here, without that it's not going >> to work. >> > > For fixed frequency, you wouldn't need bifilar. Just tune both sides > and live with a lot of leakage inductance. A few turns of HV wire on > opposite sides of a little Micrometals toroid maybe. >
Tuning is very much poo-poo'ed on in a production environment. That's why I use bifilar even for fixed frequency clocks. And it's not problem, since you have to wind the transformer anyhow why not to it bifilar?
> How about a couple of Mini-Circuits RF transformers, balanced on both > ends, connected with a pair of low capacitance, high-voltage ceramic > caps? >
Radio Yerevan would say: Basically yes, but ... in a medical environment you have pretty strict UL60601 leakage current requirements. Microamp range and even when measured at 60Hz you can quickly reach the limit. Usually because this 200MHz clock isn't the only thing that needs to cross the barrier. There's typically also the need to send a few watts of juice across. Then there is the defibrillator jolt itself. 5kV out of a 32uF cap. A mishap could send personnel into cardiac arrest. Especially when the paddles don't find the low impedance path into the patient's chest for some reason. The main reason why such a series connection is tough and will likely be shot down at agency approvals is that there are no real guarantees how much voltage the Mini-Circuits XFMR and how much the high voltage cap will get. Unless you have bleeders across each or ditch the Mini-Cicuits part (because you don't really need it, might as well use an LVDS receiver). -- Regards, Joerg http://www.analogconsultants.com/
On Nov 8, 2:32=A0pm, Joerg <inva...@invalid.invalid> wrote:
> John Larkin wrote: > > On Tue, 08 Nov 2011 10:55:39 -0800, Joerg <inva...@invalid.invalid> > > wrote: > > >> halong wrote: > >>> Hi all, > > >>> I have a 200 MHz clock signal to send across the barrier of 5 KV > >>> I'm thinking of some RF transformer - perhaps "air core" ... but can'=
t
> >>> find any good candidate > > >> Medical, defibrillator-proof? That's pretty much my home turf. I often > >> use ferrite toroids, mostly #43 material and sometimes #61. Then 2-3 > >> turns, bifilar winding. Bifilar is key here, without that it's not goi=
ng
> >> to work. > > > For fixed frequency, you wouldn't need bifilar. Just tune both sides > > and live with a lot of leakage inductance. A few turns of HV wire on > > opposite sides of a little Micrometals toroid maybe. > > Tuning is very much poo-poo'ed on in a production environment. That's > why I use bifilar even for fixed frequency clocks. And it's not problem, > since you have to wind the transformer anyhow why not to it bifilar? > > > How about a couple of Mini-Circuits RF transformers, balanced on both > > ends, connected with a pair of low capacitance, high-voltage ceramic > > caps? > > Radio Yerevan would say: Basically yes, but ... in a medical environment > you have pretty strict UL60601 leakage current requirements. Microamp > range and even when measured at 60Hz you can quickly reach the limit. > Usually because this 200MHz clock isn't the only thing that needs to > cross the barrier. There's typically also the need to send a few watts > of juice across. > > Then there is the defibrillator jolt itself. 5kV out of a 32uF cap. A > mishap could send personnel into cardiac arrest. Especially when the > paddles don't find the low impedance path into the patient's chest for > some reason. > > The main reason why such a series connection is tough and will likely be > shot down at agency approvals is that there are no real guarantees how > much voltage the Mini-Circuits XFMR and how much the high voltage cap > will get. Unless you have bleeders across each or ditch the Mini-Cicuits > part (because you don't really need it, might as well use an LVDS receive=
r).
> > -- > Regards, Joerg > > http://www.analogconsultants.com/
Bifilar coils sounds promising, unless I have to find such bifilar with 5KV insulation... Another thing is how to pass the EMI regulation... how about bifilar inside a Faraday cage ? but how can I split and isolate the cage now ? Right now I'm leaning to this: "divide the clock 1/2, or even 1/4, send it over the above Analog device, then use PLL to get it back on the other side"... how's that? Thanks,
On Nov 8, 3:41=A0pm, halong <cco...@netscape.net> wrote:
> Hi all, > > I have a 200 MHz clock signal to send across the barrier of 5 KV > I'm thinking of some RF transformer - perhaps "air core" ... but can't > find any good candidate
You want a transmission line transformer, made with coaxial cable that can stand 5kV between inner and braid. Many can - Farnell lists an RG58 cable good for 15kV (5mm diameter) and an RG174A got to 6kV (2.54mm O.D.). -- Bill Sloman, Nijmegen
halong wrote:
> On Nov 8, 2:32 pm, Joerg <inva...@invalid.invalid> wrote: >> John Larkin wrote: >>> On Tue, 08 Nov 2011 10:55:39 -0800, Joerg <inva...@invalid.invalid> >>> wrote: >>>> halong wrote: >>>>> Hi all, >>>>> I have a 200 MHz clock signal to send across the barrier of 5 KV >>>>> I'm thinking of some RF transformer - perhaps "air core" ... but can't >>>>> find any good candidate >>>> Medical, defibrillator-proof? That's pretty much my home turf. I often >>>> use ferrite toroids, mostly #43 material and sometimes #61. Then 2-3 >>>> turns, bifilar winding. Bifilar is key here, without that it's not going >>>> to work. >>> For fixed frequency, you wouldn't need bifilar. Just tune both sides >>> and live with a lot of leakage inductance. A few turns of HV wire on >>> opposite sides of a little Micrometals toroid maybe. >> Tuning is very much poo-poo'ed on in a production environment. That's >> why I use bifilar even for fixed frequency clocks. And it's not problem, >> since you have to wind the transformer anyhow why not to it bifilar? >> >>> How about a couple of Mini-Circuits RF transformers, balanced on both >>> ends, connected with a pair of low capacitance, high-voltage ceramic >>> caps? >> Radio Yerevan would say: Basically yes, but ... in a medical environment >> you have pretty strict UL60601 leakage current requirements. Microamp >> range and even when measured at 60Hz you can quickly reach the limit. >> Usually because this 200MHz clock isn't the only thing that needs to >> cross the barrier. There's typically also the need to send a few watts >> of juice across. >> >> Then there is the defibrillator jolt itself. 5kV out of a 32uF cap. A >> mishap could send personnel into cardiac arrest. Especially when the >> paddles don't find the low impedance path into the patient's chest for >> some reason. >> >> The main reason why such a series connection is tough and will likely be >> shot down at agency approvals is that there are no real guarantees how >> much voltage the Mini-Circuits XFMR and how much the high voltage cap >> will get. Unless you have bleeders across each or ditch the Mini-Cicuits >> part (because you don't really need it, might as well use an LVDS receiver). >> >> -- >> Regards, Joerg >> >> http://www.analogconsultants.com/ > > Bifilar coils sounds promising, unless I have to find such bifilar > with 5KV insulation...
As I said, got to buy certified wire for that. There is no way around MIL-STD 22759 here.
> ... Another thing is how to pass the EMI > regulation... how about bifilar inside a Faraday cage ? but how can I > split and isolate the cage now ? >
I never needed a Faraday cage even though some of my patient interface designs are housed in ABS plastics. But Faraday is easy as well: Make the bifilar transformer, twist the MIL-STD 22759 wire, leave it about 1" long. Place the transformer on the system side ground, put the shield around it (no sharp edges). Make sure the shield never cuts the twisted wire. Run the twisted wire across the barrier, and that's it. MIL-STD 22759 wire if teflon-insulated and stranded (which it usually is) tends to resist twisting so you'll have to secure that with something. The wire will want to unravel the twists. Whatever you use, make sure it has the required insulating properties. No shrink tubing from dubious sources and stuff like that.
> Right now I'm leaning to this: "divide the clock 1/2, or even 1/4, > send it over the above Analog device, then use PLL to get it back on > the other side"... how's that? >
Depends on how clean the clock needs to be. It'll also add cost, complexity, size. But it does save you from having to use a custom-made part. In my cases most clocks had to be super clean and I didn't want to take chances. Plus there was also a power transfer which required a custom transformer anyhow, and whether peoduction makes one per unit or five doesn't matter much. If you go that route _first_ make sure that the AD part has the right paperwork. For example, if you are building to category CF it must support that. IME manufacturers often wave oft or their folks get that deer in the headlight look if I mention CF. I guess they smell lawsuits or whatever :-) -- Regards, Joerg http://www.analogconsultants.com/